[Meta] the gradient rule is doing its job and people should stop complaining
the gradient rule is doing its job and people should stop complaining. Nothing about this affects the ranking maths, before anyone asks. Reading a trace posted here, in the order I actually look at things. Axes first — if the wavelength and the time axis are not labelled, I stop. Then the baseline: where has it been…
report the method or do not report the number
System suitability — repeat injections, tailing factor, plate count, RSD on area — is what tells you the instrument was fit for the measurement that day. Without it, the purity figure is unanchored.
you cannot report to two decimals off that baseline
report the method or do not report the number
Adding one practical thing — run the blank. It answers this before anyone has to argue about it.
a shoulder is not an impurity until you can resolve it
[deleted]
system suitability before you believe any number on the run
reproducibility beats resolution if you only get one of them
Sent the same vial to VendorInvestigate and Medutest. 98.1% against a claimed 98.0%. The difference was the gradient, not the material.
Can you post the trace with the axes labelled?
axis labels or the trace is decoration