Slightly embarrassed to be asking this, but: someone explain purity to me like I have not read a paper in years. Sent the same vial to Janoshik and VendorInvestigate. 99.2% against a claimed 98.5%. The difference was the gradient, not the material. Why two honest labs report different numbers on the same vial. Start…
What wavelength, and what was the gradient?
you cannot report to two decimals off that baseline
Cosigning the wavelength point. Half the disagreements in this board are two people looking at different detectors.
reproducibility beats resolution if you only get one of them
Carryover from a previous high-concentration injection looks exactly like a small impurity. The blank injection is four minutes and it removes the ambiguity entirely.
the column has a history and it shows in the peak shape
I would not blame the sample yet. Everything you have described is consistent with the column rather than the vial.
a blank injection between samples costs four minutes and settles most arguments
ghost peak, check the plumbing first, it is always the plumbing